    Default Industrial scientific itx multi-gas meter question (ds2)

    So we have a BUNCH of these DS2 docks with matching ITX (Multi-Gas) Meters from Industrial Scientific.

    Per bottle of gas we have 5 on the daisy chain... we have two daisy chains.


    How often should we be performing a bump test and calibration?

    Currently we are doing daily bump tests, and monthly calibrations.

    Is this appropriate?

    It seems as though we are running through the test gas in the two main chains every 2 weeks or so..

    Our setup looks something like this:

    TEST BOTTLE 1 --> ITX 1 --> ITX 2 --> ITX 3 --> ITX 4
    TEST BOTTLE 2 --> ITX 5 --> ITX 6 --> ITX 7 --> ITX 8
    TEST BOTTLE 3 --> BACKUP ITX 1
    TEST BOTTLE 4 --> BACKUP ITX 2

    This is our docking station:
    http://www.indsci.com/DS2/

    This is our multi gas meter:
    http://www.indsci.com/iTX/


  2. #2
    MembersZone Subscriber
    Join Date
    Mar 2001
    Location
    Edmonton, Alberta, Canada
    Posts
    15

    Default

    Hi,
    Using the docking station to bump test daily and then calibrate monthly is what the manufacturer recommends. The docking station should notify you if a bump test has failed. That detector should then be calibrated.
    I know that the docking stations tend to use a bit more gas than doing it all manually, but it sounds like you are using more than usual.
    Any leaks?
    Is the DS2 set to force calibration every time it bumps?
    Is it set to auto-calibrate if bump testing fails?
    What size cylinders?
    These are just some ideas. You could also talk to your distributor or just cal ISC directly.

  3. #3
    Forum Member
    Join Date
    May 2009
    Location
    Livonia, MI
    Posts
    35

    Default Size calgas cylinder?

    What size calibration gas cylinder are you using? If you're actually following the recommended procedure, it's going to eat up some serious gas. However, you sure be given a standing ovation for actually following the manufacturer recommendations here.

    I'd recommend going with a 58L cylinder, at least. With how much you're running, you may want to look into going with a large reusable cylinder from a company like Airgas. One of my customers, a process piping firm, who does a lot of on site work uses these and are ecstatic with how much they're saving over the disposables. You'd need a CGA regulator and all to set it up, but the guy from Airgas (or Praxair or whoever is in your area) would be able to help you out there, and if they can't I'd walk you through what you needed to do to set it all up, it's pretty easy. I'm attaching a picture of one so you'll know what to ask a rep for if you decide to go that route. They call them "30 cubic footers" usually.

    If you have any questions call me, 734 560 7481, that's my direct cell.

    Best,

    James Moore, Ideal Calibrations, LLC.
    Attached Images Attached Images  
    James Moore
    http://www.idealcalibrations.com
    ----
    "It's still alarming, how do I get it to stop?"
    "Have you tried getting your *** out of the god $#@!$# building?"

Thread Information

Users Browsing this Thread

There are currently 1 users browsing this thread. (0 members and 1 guests)

Similar Threads

  1. China Gas Field Incident kills at least 191
    By NJFFSA16 in forum Fire Wire
    Replies: 1
    Last Post: 12-26-2003, 08:51 AM
  2. World Of Fire Report: 11-15-03
    By PaulBrown in forum World of Fire Daily Report
    Replies: 0
    Last Post: 11-16-2003, 06:35 PM
  3. Manatee Fla--8" Gas Line Leak and Explosion
    By captstanm1 in forum Fire Wire
    Replies: 1
    Last Post: 10-15-2003, 07:39 AM
  4. Another gas leak
    By tanker5117 in forum Fire Wire
    Replies: 1
    Last Post: 10-13-2003, 02:16 PM
  5. SOGs' for gas leaks
    By postal79 in forum Firefighters Forum
    Replies: 2
    Last Post: 03-09-2002, 10:23 AM

Posting Permissions

  • You may not post new threads
  • You may not post replies
  • You may not post attachments
  • You may not edit your posts